Tektronix Innovation ForumEnabling Innovation in the Digital Age
Presenter: Xin Zhang
泰克最新高速信号完整性测试--涵盖收发端及链路的完整解决方案
2
High-Speed Serial TestTrends and Implications
• 100 GbE is becoming more relevant as data centers and communications networks ask for more bandwidth
• SAS 12G is needed by data centers for efficient transport of internet traffic (YouTube, Facebook, Smart Phone, etc)
• High-Speed FPGA’s are increasing in complexity to support early designs above 20Gb/sec
• Proliferation of 10+ Gb/sec signaling in the communications network
Industry/Technology Trends
• Closed data eyes requiring new techniques for transmitter and receiver equalization
• Higher data rate signals have less margin –requires de-embedding
• Edge/Slew rate speeds are difficult to characterize
• New Jitter Separation Measurements are required
• Complex 8b/10b signaling difficult to verify in PHY
Implications
2011-Tektronix Innovation Forum
高速串行设计、测试挑战
设计 检验 一致性测试
Simulation Signal IntegrityEye and Jitter Analysis
Characterization & Validation
System IntegrationDigital Validation & Debug
Serial Data Network & Link Analysis
Data Link Analysis
Digital validation & Debug
Compliance Testing
Receiver TestDirect Synthesis
Transaction Layer
Data Link Layer
Ph
ysic
al L
ayer
LogicalSub-block
ElectricalSub-block
路径Tx +
-
+
-
+
-
+
-
Rx
仿真 信号完整性眼图和抖动分析
检定和验证
系统集成数字验证和调试
串行数据网络和链路分析
数据链路分析数字验证和调试
一致性测试
接收机测
事务层
数据链路层
逻辑子块
电气子块
3 2011-Tektronix Innovation Forum
完善的泰克串行数据测试解决方案
HO
ST
端口
发射机 / 接收机
Dev
ice
端口
发射机 / 接收机
Lane
链路
RefClk
实时频谱分析仪
PLL环路带宽信号互连测试
DSA8200采样示波器TDR模块S参数测量模型提取建模
数字验证和调试协议测试
Intel最佳测试平台TLA逻辑协议分析仪
DPO/DSA/MSO70000C4Ch实时示波器20GHz带宽100GS/s采样率DPOJET平台化测试软件
Tx信号质量, 参考时钟和一致性测试
Rx容限测试
BERTSCope最高26Gbps误码测试能力具有丰富系统调试功能
4 2011-Tektronix Innovation Forum
高速串行信号测试挑战
8Gbps速率的信号仍然在FR4板材上传输,势必增加更多的信道
损耗– 信号高频分量消失殆尽,影响到对示波器带宽的需求
– 由于信道损耗,上升时间没有像数据率成倍的增长
– Tx和Rx端芯片增加了更加负载的均衡处理
Tx端需要3阶的去加重(De-emphasis)以及全新的Pre-shoot Rx端需要CTLE、DFE弥补长链路的损耗
– 为了考虑到互操作性,Tx端接口一致性测试必须考虑到信号的传输通道
由于芯片的封装,通常探头无法直接探测到信号输出管脚– 使用反嵌技术(De-embedding)消除测试走线、夹具的影响
定义了很多新型的抖动测量项目– J2、J9、DDWPS、BUJ...
5 2011-Tektronix Innovation Forum
反嵌(De-embedding)技术
反嵌会使得高频噪声放大,因此需要带限滤波器抑制带外噪声
– 示波器带宽会影响到反嵌的效果
– 带宽限制和板材有关
成功的反嵌需要良好的电路板设计以及准确的S参数测试
– 阻抗匹配、低损耗
– 无增益、过大的谐振和陷波
5 GHz 10 GHz
5GHz Filter 10GHz Filter -> Noise amplification
6 2011-Tektronix Innovation Forum
反嵌必须在测量之前完成
考虑到信号走线、互联,仿真到达接收端的信号
仿真接收端均衡,得到芯片内部真实信号的特性
对信号进行抖动、眼图测试
分离抖动类型,提出优化意见
复杂的Tx端信号测量
Signal at TX Pins Measured Signalat TP1
Apply Sparameters Signal with ChannelEffects Removed
7 2011-Tektronix Innovation Forum
Link Analysis-Embed/De-embed and EQ Post processing is required on most signals before
measurements are performed– Compliance Channels are embedded and eye is opened with a reference
equalizer as defined in the specification
Virtual probe access at intermediate test points is critical forsystem visibility
– Visually see and measure the effects of de-embedding, embedding, and equalization
8 2011-Tektronix Innovation Forum
9
Transmitter Jitter Measurements
Necessary to take transmitter jitter measurements with all lanesoperating in order to capture crosstalk effects
Measurements are taken at TP1 and de-embedded back to the pins of the TX
Necessary to separate uncorrelated and data dependent jitter in order to ensure that jitter that can be recovered is not budgeted as uncorrelated jitter
Jitter measurements Data Dependent Jitter Uncorrelated Jitter
Cause Due to package loss and reflections (dynamics in the channel, ISI)
Uncorrelated - PLL jitter, crosstalk, noise conversion (amplitude to phase)
How to compensate Can be reduced by equalization
Difficult to remove (better components, layout)
2011-Tektronix Innovation Forum
10
Transmitter Jitter Measurements: Data Dependent JitterDDJ Measurement Process Measurement taken on multiple repeats of the compliance pattern using a 1st
order CDR function representing a high pass filter
A PDF is created for each edge crossing of the compliance pattern
DDJ is calculated as the difference of the mean of each PDF and the recovered clock edge
Measurement is defined as the absolute value of DDJ(max) – DDJ(min)
2011-Tektronix Innovation Forum
11
Uncorrelated Jitter ExampleTTX-UTJ / TTX-UDJDD
• DDJ is removed from the PDF of each edge
• Data is converted to Q-Scale
• Uncorrelated Deterministic Jitter Dual Dirac (UDJDD)
– Accounts for Periodic Jitter and Crosstalk Convert the PDF to Q-Scale
• Random Jitter is implied by subtracting UDJDD from UTJ
2011-Tektronix Innovation Forum
New Jitter Separation Measurements Next Generation Standards require separate jitter budgets for Data
Dependent(DDJ) and Uncorrelated Jitter– TX and RX Equalization only compensate for DDJ
Crosstalk is a key contributor to Bounded Uncorrelated Jitter (BUJ) and is worse in multi-lane topologies
– BUJ, if counted as RJ will artificially inflate TJ – NEW! Tektronix Sampling and Real-Time platforms provide BUJ
measurements which provides more accurate jitter separation to support next generation standards
12 2011-Tektronix Innovation Forum12
Jitter Analysis AdvancesBUJ in Thunderbolt example
Tektronix Technology Innovation Forum 2011
Cloud Computing
TJ@BER1, Math1 10.105psRJ1, Math1 506.04fsPJ1, Math1 3.6968psDJ1, Math1 3.6968psNPJ1, Math1 881.89fsTIE2, Math1 55.789fsRise Slew Rate1, Math1 9.2627V/ns
TJ@BER1, Math1 9.9087psRJ1, Math1 556.41fsPJ1, Math1 2.6685psDJ1, Math1 2.6685psNPJ1, Math1 592.92fsTIE2, Math1 89.108fsRise Slew Rate1, Math1 9.2542V/ns
TJ@BER1, Math1 10.315psRJ1, Math1 680.95fsPJ1, Math1 1.7365psDJ1, Math1 1.7365psTIE2, Math1 44.029fsRise Slew Rate1, Math1 9.3228V/ns
TJ@BER1, Math1 11.159psRJ1, Math1 694.31fsPJ1, Math1 2.8264psDJ1, Math1 2.8264psTIE2, Math1 -25.694fsRise Slew Rate1, Math1 9.2843V/ns
Legacy DecompositionNew BUJ Decomposition
13
High-speed Tx testing require PriciseClock Data Recovery
1. Clock recovery is also required for many test instruments which characterize serial data systems.– Recovering clock may be desirable even when Tx
clock is available.
2. Instruments which sample the data contiguously, such are ‘real time’ oscilloscopes, implement software based clock recovery.– Generally limited to lower data rates / limited
measurement depth.
3. Sampled data instruments require a hardware clock, derived from hardware clock recovery.– Sampling oscilloscopes – for measuring higher data
rates.– Analyzing BERTs – for greater measurement depth or
higher data rates.
4. “Instrumentation Grade” clock recovery is required.– Calibrated, repeatable, and adjustable.
Decision
CircuitData In
Clock
Regenerated Data Out
Jittered incoming data
Jittered incoming data
What does CDR really do?
Regenerated data cleaned upRegenerated data cleaned up
Recovered clock moveswith the data.Recovered clock moveswith the data.
Receiver decisions relative
Receiver decisions relative
Tracking of jitter is a key attribute of clock recovery
Tracking of jitter is a key attribute of clock recovery
1.1.
2.2.
3.3.
4.4.
5.5.
CDR effect jitter measurement dramatically
10 kHz 100 kHz 1 MHz 10 MHz
Jitter Modulation Frequency, Hz
Jitter In:Jitter In:
PLL not able to track as
effectively. Jitter seen is attenuated compared to the
input.
PLL hardly able to
track input jitter at all. Output is
stable signal.
PLL tracks data closely,
jitter on output
similar to input
RecoveredClock
Jitter Out:
RecoveredClock
Jitter Out:
0
-40
-20
-60Jitte
r Tra
nsfe
r Fun
ctio
n (d
B)
= 20
Log
(Jitt
er O
ut/ J
itter
In)
1.1.
4.4.5.5.
6.6.
2.2.3.3.
Jitter Mod. Freq., Hz
Constant Amplitude SJChange Peaking by 3 dB
In: 35ps Jitter
1.
Out: 35ps Jitter
2.
Out: 50ps Jitter
3.
The Effect of Peaking Illustrated
1 MHz loop bandwidth, 400 kHz SJ modulation
The eye diagram from the green response shows jitter gain – more jitter out of the device than was present on the input.
The eye diagram from the green response shows jitter gain – more jitter out of the device than was present on the input. 4.4.
The Effect of Edge Density
1. Different data patterns have different “edge density” or “transition density”.– A 1010 clock pattern has a 100 % edge
density.– A true PRBS pattern has a 50 % edge density.
LBW2. The pattern’s edge density effects the
energy entering the PLL phase detector, which affects the loop response, loop bandwidth.
3. The calibration of loop response requires knowledge of the edge density.
Measurements: “Instrumentation Grade” Clock Recovery Units
1. Clock Recovery loop response affects the amount of jitter “seen” by serial data analysis instruments.– Loop bandwidth– Roll off slope– Peaking
2. The parameters must be tightly controlled for accurate and repeatable jitter measurements.
3. The ability to adjust these parameters is often required.– CR characteristics are often dictated by many Compliance Test
Standards.– Setting the parameters to match those in receiver allows the instrument
to “see” the data as receiver does in the real system.
Visual Trigger and Serial Decode
Next generation designs have less margin and additional analysis must be done to pinpoint in on pattern dependent issues
– NEW! Visual Trigger qualifies hard to define trigger events 8 customizable shapes for capture of real
signal behaviors
Electrical and Logic layer are merging and requires simultaneous analog and protocol views
– NEW! 8b/10b Serial Decode Trigger or Search on decoded traffic Compare to analog views to speed up time to
answer
2011-Tektronix Innovation Forum20
High Speed Device connectivity Solder-Down probing is required when a signal needs
to be measured and no SMA or RF connector is available
High Speed standards like PCI Express are commonly seen in embedded designs
Debugging often requires signal access as close to the TX and RX as possible
TriMode probing enables simultaneous access of single ended and differential signals– No need to re-solder probe tips to switch between
measurements (i.e. common mode measurements)
World’s Highest Performing Solution For Next Generation Serial Technologies21 2011-Tektronix Innovation Forum
信号产生
高速串行信号不断增加的数据率
不断增加的信号带宽
复现现实世界中的信号
复现传输线效应
产生各种干扰信号:抖动、噪声和其他干扰,并精确的知道干扰量的大小
高速串行系统Rx端容限测试
pathTx +
-
+
-
+
-
+
-
Rcv
单纯考察Tx端是不足以保证BER
Rx端复杂的结构:Equalizer、CDR
没有Tx和传输链路时对Rx的压力测试
对Rx端施加各种各样的压力和一致性测试信号
22 2011-Tektronix Innovation Forum
Why Receiver Testing is Different
ReceiverBright…Bright… Shiny…Shiny…
New…New…
“Receiver”, “Re-Timer”“Decision Circuit”, “SERDES”
Poor quality bits in…. Pristine bits out…..
• Transmitters are tested with eye diagram analysis but a Receiver Changes Everything
• Can no longer rely on how good the eye looks as a measure of performance….….the eye shape only tells how nice the output stage is.
Receivers are Tested with BER
Receivers
What causes “poor bits”?
无论是针对设计还是生产制造,规范都定义了明确的Rx端测
试需求
所有标准都要求进行Jitter Tolerance 一致性测试
被测设备类型:– SerDes – Transceivers – Multi Media Sink devices– Rx devices
各种高速串行总线对Rx接收端测试的要求
Standard Data Rate
Jitter Tolerance
Timing Skew
Amplitude Sensitivity Emphasis
SATA Gen 3 6 Gb/s - -PCI Express 2.0 5 Gb/s
PCI Express 3.0 8 Gb/s
HDMI 1.40.75 Gb/s
to 3.4 Gb/s
-
FC 4, 8 G4.25 Gb/s
to 8.5 Gb/s
DisplayPort2.7 Gb/s
5.4Gb/s
USB 3.0 5 Gb/s -
25 2011-Tektronix Innovation Forum
各种高速串行总线对Rx接收端测试的要求
1.设置DUT进入Loopback模式(Analog/Re-timing)
2.产生规范要求的抖动分量,在不同的频点上分别产生相应的抖动量
3.将stressed信号注入DUT Rx4.统计DUT Tx端发出的信号的误码率是否达到要求
12
4
3
26 2011-Tektronix Innovation Forum
BERTScope产生各种压力类型
Sine jitter– 1KHz~100MHz– max.1100ps
Random jitter– f>1GHz
Bounded PRBS jitter
SSC– 12,500ppm
Sine Interference
27 2011-Tektronix Innovation Forum
动态改变速率、压力、码型等
6 Gb/sPRBS-7DJ:SJ:RJ:SI:
6 Gb/sPRBS-7DJ:SJ:RJ:SI:
6 Gb/sPRBS-7DJ:SJ:RJ:SI:
11
22
33
28 2011-Tektronix Innovation Forum
PCIe Gen 3 Stress Recipe - Overview
Tx Eq
8G PRBSGen
RJSource
SJSource
Combiner
DiffInterference
Cal.Channel
ReplicaChannel
Test Equipment
CMInterference
Post‐processing
Eye HeightAdjust
(Taken from PCI Express Base Spec, Figure 4‐71)
PCI Express Gen 3 uses a long circuit board channel that closes the eye, and two forms of vertical eye closure (‘Interference’).
PCI Express Gen 3 uses a long circuit board channel that closes the eye, and two forms of vertical eye closure (‘Interference’).
29
PCIe Gen 3: Example Add-In Card Stress Calibration
In Out+
‐
+
‐
To RT Scope for calibration
SI Combiner
Gen 3 CBB Riser
Gen 2 CLB
Gen 3 CBB (Main)
Rx Lane 0
Tx Lane 0
Last Cal. details being refined. This setup beingsuccessfully used at Plugfests
Last Cal. details being refined. This setup beingsuccessfully used at Plugfests
30
自动化Jitter Tolerance一致性测试方案
Test for compliance
Use Search mode to find device limits
PCIE GEN1,2,3
SATA I,II,III
USB3
Display Port
XFP/XFI
10GBase-KR
Optical
Serial Bus...
31 2011-Tektronix Innovation Forum
完整的高速串行系统测试、调试过程
测量BER测量BER
BER是否合格BER是否合格
测量眼图测量眼图
眼图是否闭合眼图是否闭合
测量BER等高线测量BER等高线
定位故障点定位故障点
分析误码率事件规律
分析误码率事件规律
Pattern SensitivityPattern Sensitivity
Error Free IntervalError Free Interval
BER Strip ChartBER Strip Chart
CorrelationCorrelation
PassPass
检查Tx端预加重
连接可靠增加驱动
检查Tx端预加重
连接可靠增加驱动
Y
Y
N
N
根据误码事件特点找到误码根源解决误码问题
根据误码事件特点找到误码根源解决误码问题
32 2011-Tektronix Innovation Forum
参考时钟泄露
多余的波形 :出现的频度不高
内存芯片设计
33 2011-Tektronix Innovation Forum
深入调试“异常”现象
将BERTScope的采样器放置在“可疑”区域
BERTScope将统计落在采样器上的码型出现的规律– 可以使用Error Correlation
出现非常有规律的事件,间隔为“24”34 2011-Tektronix Innovation Forum
解决问题
推论:Data Rate的24 分之一为时钟的频率,因此,可以定位到
异常的眼图有可能是时钟串扰引起的。
行动:增加时钟和数据走线之间的隔离;减缓时钟的slew rate
35 2011-Tektronix Innovation Forum
22-SEPT-201036
全新的“逻辑协议测试”理念
示波器 逻辑协议分析仪
Logic Protocol Analyzer for PCI Express 3.0Performance You Can See - One tool to span from protocol to physical
协议分析仪
Summary Profile WindowTransaction Window
? Poor physical layer support
OpenEYEFastSYNCScopePHY
36 2011-Tektronix Innovation Forum
22-SEPT-201037
TLA7000平台+逻辑协议分析模块
支持高速串行总线协议– PCIE Gen1/2/3– DDR2/3– USB3– SATA/SAS– Serial RapidIO– FC...
对信号模拟特性采集
– 可以无缝的和示波器联合使用
– 时间相关的模拟、数字、协议层联合测试
– 不漏失任意一个总线异常
– 直接定位故障到系统模拟层面
灵活多样的探测方式
– 焊接式、嵌入式以及自定义信号探测方式,支持最多样的总线拓扑
– 内置均衡处理模块,真实可靠反映信号模拟特征
TCA-SMA TekConnect Adapters shown connected to 1-lane (2 differential inputs) out of the 2-lanes per P67SAxxx probe connector. Unconnected lanes have 50ohm terminators connected at end of coax.
TLA70122-module
logic analyzer
mainframe2x TLA7SA16
x8 PCIe3Modules
DPO/DSA70000B
4-channel33 GHz
oscilloscopewith
DPOJETloaded withP67SAxxx
S-parameter file to
de-embed the probe
P67UHDSMA2-lane probe leadset
37 2011-Tektronix Innovation Forum
TEKTRONIX CONFIDENTIAL
Information Density - Transaction Window (1/3)
Packet pane to view fields (can simultaneously open multiple
packets
Toolbar(Search, filter,
display management)
Status bar &access to “Summary
Profile Window”
BEV(Bird’s Eye
View)
ConfigLinks
Transaction stitching shows
packets participating in transaction – or
incomplete transactions as errors – mouse over shows time
–arrowheads/squares filled or not
based upon completion
PHY layer info (shows sub-packet info
such as ordered sets –view in more
detail in Listing Window)
Flow control credit
tracking
Each row of the Packet View pane
represents a single packet
22-SEPT-201038 Transaction - Link based behavior of protocol elements (transactions, packets, fields, ordered sets)38 2011-Tektronix Innovation Forum
22-SEPT-201039
协议逻辑分析仪--协议层、模拟层联合观测
Waveform symbolic decode
(lane alignment
disabled for accurate time correlation)
Waveform - Time based view of the data on each lane
39 2011-Tektronix Innovation Forum
Tektronix高速计算机验证系统
DDR3 Validation DIMM *
PCI Express 3
DDR3 Interposer Probe *
CPU Bus
Electrical compliance test
DD
RD
DR
DD
R
IOH
CPU
DD
R
DD
R
DD
R
DD
R
DD
R
DD
R
PCIe3SATA
Other buses
* Nexus Technology and FuturePlus products available for DDR2 and DDR3 including 1600MT/s
*Courtesy of Intel.
业内性能最高的TLA7BB4 模块,轻松应对DDR3-1867逻辑协议测试
定时采样率50G/s
状态速率3GHz
2011-Tektronix Innovation Forum
Jan 2011 Tektronix最新测试测量解决方案构架
全新的信号完整性调试理念-模拟数字联合调试
TriggerState
Machine
4 ch
CH 1
CH 2
CH 3
CH 4
DPODPOLALA
Analog In
CH 1
CH 2
CH 3
CH 4
Analog Out2 GHz
AnalogMux
34 ch
34 ch
34 ch
34 ch
单LA探头连接信号,同时测量
信号模拟、逻辑特性
LA探头所有通道模拟带宽指标2~3 GHz
LA中136通道任意4路可以输出
到外部示波器
新
Jan 2011 Tektronix最新测试测量解决方案构架
模拟、数字联合调试方案
毛刺捕获技术
– 实时动态监控信号中的异常逻辑
– 高亮标注
– 准确定位异常发生时刻、位置
iView– 将逻辑分析仪和示波器无缝连接
为一套测试系统
– 在逻辑分析仪屏幕上显示自动时间对齐后的同一信号模拟和数字的波形
– 通过逻辑分析仪对示波器的触发控制,准确定位信号异常时的模拟采集
Jan 2011 Tektronix最新测试测量解决方案构架
数字、模拟联合调试系统
TLA Logic Analyzer P6860 Probe (34ch)
TLA BNC Cables
iCapture: 逻辑分析仪集成式数字/模拟探头
TLA iViewTLA逻辑分析仪 DPO/DSA示波器
ZhangXin,Tektronix China44
iView 外部示波器接口电缆
USB接口
USB-to-GPIB接口
连接至TLA7000逻辑分析仪 连接至DPO示波器
小结
业内最丰富的针对于高速计算机系统
测试方案,从链路层、模拟层到逻辑协议层
业内最高的误码率分析仪,提供了深邃的误码率事件分析功能;一键式接收端容限测试
业内最新的“逻辑协议”分析仪,结合
了模拟层、数据层和协议层的调试、测试于一体
支持广泛的高速总线标准测试
45 2011-Tektronix Innovation Forum
Link Analysis Post processing is required on most signals before
measurements are performed– Compliance Channels are embedded and eye is opened with a reference
equalizer as defined in the specification
Virtual probe access at intermediate test points is critical forsystem visibility
– Visually see and measure the effects of de-embedding, embedding, and equalization
46 2011-Tektronix Innovation Forum
New Jitter Separation Measurements Next Generation Standards require separate jitter budgets for Data
Dependent(DDJ) and Uncorrelated Jitter– TX and RX Equalization only compensate for DDJ
Crosstalk is a key contributor to Bounded Uncorrelated Jitter (BUJ) and is worse in multi-lane topologies
– BUJ, if counted as RJ will artificially inflate TJ – NEW! Tektronix Sampling and Real-Time platforms provide BUJ
measurements which provides more accurate jitter separation to support next generation standards
4747 2011-Tektronix Innovation Forum
Visual Trigger and Serial Decode
Next generation designs have less margin and additional analysis must be done to pinpoint in on pattern dependent issues
– NEW! Visual Trigger qualifies hard to define trigger events 8 customizable shapes for capture of real
signal behaviors
Electrical and Logic layer are merging and requires simultaneous analog and protocol views
– NEW! 8b/10b Serial Decode Trigger or Search on decoded traffic Compare to analog views to speed up time to
answer
48 2011-Tektronix Innovation Forum
Bandwidth Impact on Measurement Accuracy
An Example, SAS 12G
At 12G, only the entire First Harmonic is captured– Edge definition comes after the 2nd Harmonic– Resulting signal is a sine wave and results in higher TJ and an eye
height that is larger than expected when baselined with a sampling scope (470mV)
49 2011-Tektronix Innovation Forum
Bandwidth Impact on Measurement Accuracy
At 24G only the entire 3rd harmonic is captured– Note the peanut in the center of the eye that closes the eye– TJ is improved, but eye height is lower than the sampling scope baseline
50 2011-Tektronix Innovation Forum
Bandwidth Impact on Measurement Accuracy
At 33G the entire 5th harmonic is captured– TJ is reduced and eye height correlates with the sampling scope
51 2011-Tektronix Innovation Forum
Receiver Testing is Required for Certification
Compliance Testing is just the start of characterizing RX performance
Margin Testing can lead to insight on system margins and errors that can cause interoperability problems
BERTScope enables insight into RX failures Eye diagram for quick
diagnosis of synchronization and BER failure issues
Debug challenging signal integrity problems
Error Location Analysis
Pattern Capture
Jitter Map
BER Contour
JitterJitter
Error Correlation
Error Correlation
BERBER
Jitter DecompositionJitter Decomposition
Jitter ToleranceJitter Tolerance
PLUS…PLUS…
52
Proven Expertise in High Speed Serial Data Testing Transmitter Signal Quality
4Ch 33Ghz Real Time Oscilloscopes High Bandwidth Differential Probes Jitter and Eye Analysis Serial Data Link Analysis
Digital Validation and Debug Protocol and Logic Analyzers Protocol Debug and Validation
Receiver Testing BERTScope Bit Error Rate Teste Jitter Tolerance Test
S Parameters
Interconnect Testing DSA8300 Sampling Oscilloscopes TDR Heads TDNA SW
53 2011-Tektronix Innovation Forum